[Diabetes-Talk] Issues With Medtronic Insulin Pumps.

Patricia Maddix pmaddix at comcast.net
Mon Feb 16 22:01:17 UTC 2026


This is indeed very interesting and very unfortunate. I see that the situation also applies to the most current 780 pump as well as older ones and I may have encountered these issues way back when I was using my paradigm pump up until 2022.
Patricia
Sent from my iPhone

> On Feb 16, 2026, at 11:20 AM, Eileen Scrivani via Diabetes-Talk <diabetes-talk at nfbnet.org> wrote:
> 
> Even though I no longer use a Medtronic pump, some of you might still have one or the 600G. I got the below notice in email today so am cutting and pasting below.
> 
> 
> 
> Medtronic logo header
> 
> URGENT MEDICAL DEVICE CORRECTION
> 
> MiniMed™ Paradigm™, 600 series, and 700 series insulin pump systems
> 
> Pump Placement with Respect to Infusion Site
> 
> 
> 
> Products Impacted
> 
> Model/CFN Number
> 
> MiniMed™ Paradigm™ Insulin Pumps (all models)
> 
> MiniMed™ Paradigm™ Insulin Pumps (all models)
> 
> MiniMed™ 630G, 670G, 770G, 780G Insulin Pumps (all models)
> 
> 
> 
> Please visit
> 
> this page
> 
> for affected user guide versions.
> 
> 
> 
> February 2026
> 
> Medtronic reference: FA1514
> 
> 
> 
> Dear Valued Customer,
> 
> 
> 
> We are contacting you with guidance about how to position your pump and infusion set. Please note that this issue and the required actions apply to all
> 
> MiniMed™ Paradigm™, MiniMed™ 600 Series, and MiniMed™ 700 Series pump models.
> 
> 
> 
> Issue Description
> 
>  Due to environmental conditions, there may be slight variability in the amount of insulin delivered when your pump is worn at distances above or below
> 
> the location of your infusion set.
> 
>  This variability may increase with increasing distances above or below your infusion set location.
> 
>  Placing the pump above your infusion site can cause over-delivery and placing it below your infusion site can cause under-delivery.
> 
>  The greatest variability in insulin delivery may occur at distances beyond 14 inches / 35.5 cm above or below your infusion set location.
> 
>  This variability in insulin delivery will not be captured in your pump’s history or CareLink reports.
> 
> Risk to Health:
> 
>  Individuals who require lower daily insulin doses and those with high insulin sensitivity may experience changes to their blood sugar (hypoglycemia or
> 
> hyperglycemia) because of this phenomenon.
> 
>  If you are unsure whether this applies to you, it is important that you seek guidance from your healthcare professional.
> 
> As of February 9, 2026, Medtronic has not identified any user complaints that can be attributed to this issue.
> 
> 
> 
> Medtronic MiniMed will update the User Guides for our currently distributed insulin pumps to include the following:
> 
> 
> 
> “It is important to keep the location of your pump stable relative to your infusion site. Do not wear or place your pump more than 14 in (35.5 cm) above
> 
> your infusion site. Doing so can cause an over-delivery of insulin, which may result in hypoglycemia.”
> 
> 
> 
> User guides that are updated will be available at:
> 
> https://www.medtronicdiabetes.com/download-library.
> 
> 
> 
> Your Required Actions:
> 
>  Keep the location of your insulin pump close to your infusion site.
> 
>  If you need to raise the pump above your infusion site, such as when a parent or caregiver is using the pump for a child, limit the time that the pump
> 
> is held above the infusion site.
> 
>  Pay attention to any alerts from your pump as well as symptoms of hypoglycemia and hyperglycemia; follow your healthcare professional's treatment instructions
> 
> in these
> 
>  If you have any concerns about how this may impact insulin delivery and blood glucose, please contact your healthcare professional. You may also contact
> 
> us at 1-855-587-4950 with any questions about this notification.
> 
> 
> 
> _______________________________________________
> Diabetes-Talk mailing list
> Diabetes-Talk at nfbnet.org
> http://nfbnet.org/mailman/listinfo/diabetes-talk_nfbnet.org
> To unsubscribe, change your list options or get your account info for Diabetes-Talk:
> http://nfbnet.org/mailman/options/diabetes-talk_nfbnet.org/pmaddix%40comcast.net



More information about the Diabetes-Talk mailing list