The Role of Human Factors in the Design and Development of an Insulin Pump
J Diabetes Sci Technol Vol 6, Issue 2, March 2012
www.journalofdst.org
261
260
Journal of Diabetes Science and Technology
Volume 6, Issue 2, March 2012
© Diabetes Technology Society
SYMPOSIUM
The Role of Human Factors in the Design and Development
of an Insulin Pump
Noel E. Schaeffer, Ph.D.
AbstractThis article discusses human factors (HF) processes and how they are applied during the development of a medical device to minimize the risk that
the user interface design could lead to patient errors, adverse events, and product recalls. This process is best defined as “prevention through design.”
The HF design process is exemplified by three distinct phases: (1) preliminary analysis, (2) formative design evaluation and modification, and (3) design
validation. Additional benefits of employing HF principles during medical device development are briefly reviewed, including reduced patient risk by eliminating
design flaws, increased patient adherence through the reduction in the complexity of therapeutic regimes, and reduced likelihood for product recalls.J
Diabetes Sci Technol 2012;6(2):260-264
Introduction
There are approximately 450,000 people in the United
States who use insulin pumps to manage their diabetes.
All these people interact with their pumps several
times every day to administer insulin into their bodies
to manage their blood glucose levels. If you have ever
interacted with a piece of technology regularly—for
example, the automated ticket machine for the train, a
smart phone, or the check-in kiosk at the airport—and
you have thought to yourself, if only the designers and
engineers had done something a little differently, it would be
so much easier to use, then you would have been thinking
in terms of how people interact with technological
systems and of how important it is (especially with
medical devices) that those interactions are consistently
safe, error free, and efficient.
Human Factors Defined
Human factors (HF) is a unique field because it draws
from a variety of disciplines, including psychology,
engineering, design, statistics, ethnography, and anthro-
pometry. Specifically, a human factor is any physical,
perceptual, cognitive, or behavioral aspect of a human
being that impacts a technological system or environment.
The most important aspect of HF is that it is truly an
empirical science, which means it is based almost entirely
on observation. The power of the HF discipline is that
its process can be wielded to collect information on how
humans interact with technology, and that knowledge
can be applied to the design and development of
technologies that better fit human behaviors, needs, and limitations. The focus of HF is expressed by the seminal
work of Sanders and McCormick:
Human factors focuses on human beings and
their interaction with products, equipment, facilities,
procedures, and environments, used in work and
everyday living. The emphasis is on human beings
(as opposed to engineering, where the emphasis is
more strictly technical engineering considerations) and
how the design of things influences people. Human
factors, then, seeks to change the things people use
and the environments in which they use these things
to better match the capabilities, limitations and needs
of people.1
A common misperception is that HF research is similar
or related to market research. Market research typically
utilizes methodologies such as focus groups, customer
preference surveys, or executive reviews to gather data that
center on opinions, attitudes, and speculation about a
product. Focus groups and other kinds of market research
methodologies are not specific tests of medical device
use or risks in medical device interaction. They provide
a different kind of information, which can be useful
during the very early stages of product development when
developing user profiles, creating device-use scenarios,
and better understanding use environments.
The Human Factors Process and Medical
Device Development
The health care field has come under major scrutiny
because of poorly designed medical systems and devices
with inherent design flaws that induce user errors.
There have been numerous examples of infusion pump
recalls due to poor designs. For example, in August 2004,
the N’Vision® Clinical Programmer was recalled by the
Food and Drug Administration (FDA) because users
were able to enter infusion times with incorrect units.
Minutes were entered into the hours field, which caused
patient overdose.2 The Colleague® IV infusion pump
could shut down while delivering critical medication to
patients. The reason for this was that the “on/off” key
was so close to the “start” key that nurses would often
inadvertently turn the pump off when they intended to
start delivery. Over 206,000 Colleague infusion pumps,
used mostly in hospitals, were recalled by the FDA.3,4
The HF process in the medical device field takes a very
risk-centric approach, generally referred to as a “prevention
through design” strategy. The primary objective is to
design out characteristics of a system or device that
could lead to human error. The HF process generally
includes three primary phases: (1) preliminary analysis,
(2) formative evaluation and design modification, and
(3) validation testing. The t:slimTM insulin delivery system
(t:slim pump) user interface was developed with this process.
This process resulted in a pump that utilizes a touch screen
to facilitate data entry, along with a home screen to
display critical information users need to manage their
diabetes. The home screen displays critical information
such as the amount of insulin in the cartridge, battery
life, time and date, and insulin on board, as well as
provides direct access to bolus functions and pump
system functions (“options”).
Preliminary analysis started prior to the design of the
t:slim and sought to better understand who the users are,
what their needs are, what kind of environment they use
the device in, what their limitations are, what specific
tasks they need to accomplish, and a risk analysis of
those tasks as they related to device use. This phase also
included analysis of the user interface designs of products
already in use. The methodologies, study focus, and
outcomes that occurred during the preliminary analysis
phase are summarized in Table 1. From the data collected
in the preliminary analysis, an early interface design
was developed, as shown in Figure 1.
The next phase of t:slim pump development involved
HF activities that included formative evaluation and
design modification. This phase of the process was
accomplished via risk-based iterative usability testing.
The primary focus of this testing was on safety and
usability of the product’s design. A typical usability study is relatively small (5–8 participants), with actual
users being observed by a researcher in a controlled
environment. The participants go through a set of task-
based scenarios that are representative of real device
use. The researcher observes each participant’s behavior
as they are going through the tasks and records data on
their interactions.5,6 Data collected are used to inform the
next iteration of the design. Table 2 shows a summary
of the formative studies that were conducted during the
development of the t:slim pump.
During several of the early usability tests, it was revealed
that participants struggled with data entry, navigation,
and information retrieval tasks. One example of design
improvements to the home screen that occurred as a
result of this iterative usability testing is exemplified
in Figure 2.
Modifications were continually made to the user interface
based on observations made during the formative design
evaluation phase. At the end of the development process
and formative tests, a HF validation test was conducted
to prove that the t:slim pump was safe and effective for
human use.
The purpose of the validation study was to confirm that
the t:slim pump could be used safely and effectively by
representative users in the intended use environment.
Study participants included both people with diabetes
who were insulin pump users and people with diabetes
being treated with multiple daily injections. Participants
performed a comprehensive, realistic set of basic and
advanced task-based scenarios with the t:slim pump.
Data were collected via a software data log, participant
journals, follow-up interviews, and video/audio tape
recordings. The results indicated that, during certain
tasks, participants made programming errors when
inputting information into the pump. To mitigate
potential programming errors, a software solution was
developed that forced users to actively confirm their
inputs before continuing through the workflow. In order
to prevent automaticity7,8 or response chaining, in which
users could act habitually in an automatic fashion and
tap through confirmation screens that were presented
too frequently, the confirmation screens appear only
when users make changes to settings that impact insulin
delivery, such as delivering a bolus or setting a basal
rate. A revalidation study was conducted to evaluate the
effectiveness of the active confirmation screens. The goal
was to demonstrate and validate the effectiveness of the
active confirmation screens on the t:slim pump with a
sample of representative users.
There were several events that occurred during the
revalidation study that revealed that the active confirmation
screens did prevent incorrect entries. The reason the active
confirmation screens may have impacted participant
behavior is likely due to several related factors, including
the unique screen layout, a summary of all critical
information in one location, and the active confirmation
required from the participant. The home screen of the
validated design is shown in Figure 3.
The improvement in the usability of the interface was
further substantiated through a system usability scale
(SUS) questionnaire that participants completed after their
sessions. The SUS yields a single number which represents
the overall usability of the system.9 Figure 4 shows
higher SUS scores for the new interface versus the old.
The results of these validation studies were the culmination
of successful HF/usability design efforts, which included
preliminary analysis, iterative formative evaluations, and
design modifications. The current regulatory environment
has required an increased focus on the application of
the HF process to the development of medical devices in
order to eliminate design flaws that could lead to unsafe
behavior. These studies exemplified how applying that
process helped create a user-centric system that reduced
risk, increased effectiveness, and improved ease of use.
Human Factors Standards
There are several HF standards that practitioners and
companies can use to guide them through the practical
application of HF processes. The Association for the
Advancement of Medical Instrumentation is an organization
that brings together experts in order to create standards that are widely recognized by other organizations, including the
FDA. Many of these documents contain comprehensive
information that covers HF design processes as well as
HF design principles.10–12
Conclusion
There are several primary benefits to employing a HF
process during the design of medical devices. The most
important benefit is increased patient safety. This safety
comes in the form of reduced patient risk through a
design that has been heavily tested with representative
users in its intended use environment in order to
eliminate design flaws. Another benefit may be increased
patient adherence. There is evidence indicating that reducing
the complexity of therapeutic regimes may increase
patient adherence.13–16 Reduced therapeutic complexity,
in part, can be derived from improved device usability.
Secondary benefits of a systematic HF process may include
reducing or eliminating the need for costly modification(s)
after product launch, reducing the likelihood of product
recalls due to design flaws, improving overall ease of
use (improved ease of use is a byproduct of good HF
practices, not the sole outcome), and enhancing look and
feel. There is also evidence to suggest that patients may
pay a higher premium for more usable products.17 All of
the benefits resulting from the systematic implementation
of a sound HF process will converge to generate a
medical product that creates a patient experience that is
safer, more efficient, and more satisfactory.
2041
Funding:
This work was funded by Tandem Diabetes Care Inc., San Diego, CA.
2058
Disclosure:
Noel E. Schaeffer, Ph.D., is a full-time employee at Tandem Diabetes
Care.
2077
Acknowledgments:
I thank all the HF practitioners and user-experience designers who
contributed their time, energy, and expertise to the development of
the t:slim pump, including Kathryn Rieger-King, Ph.D., Brian Bureson,
and Jason Farnan.
N’Vision® is a registered trademark of Medtronic Inc. Colleague® is a
registered trademark of Baxter Healthcare Corp.
2139
References:
1. Sanders MS, McCormick EJ. Human factors in engineering and
design. 7th ed. New York: McGraw-Hill; 1993.
2. Food and Drug Administration. Medtronic 8870 Software
Application Card Version AAA 02: class I recall. http://www.fda.
gov/MedicalDevices/Safety/RecallsCorrectionsRemovals/ListofRecalls/
ucm064764.htm. Accessed October 31, 2011.
3. Food and Drug Administration. Baxter Healthcare Corp.
COLLEAGUE® Volumetric Infusion Pumps: class 1 recall. http://
www.fda.gov/MedicalDevices/Safety/RecallsCorrectionsRemovals/
ListofRecalls/ucm063713.htm. Accessed October 31, 2011.
4. Reuters. F.D.A. orders Baxter recall of IV pumps. New York Times;
July 14, 2010.
5. Tullis T, Albert B. Measuring the user experience: collecting,
analyzing, and presenting usability metrics. Amsterdam: Elsevier;
2008.
6. Kuniavsky M. Observing the user experience: a practitioner’s
guide to user research. San Francisco: Morgan Kaufmann; 2003.
7. Schneider W, Shiffrin RM. Controlled and automatic human
information processing. I. Detection, search, and attention. Psychol
Rev. 1977;84(1):1–66.
8. Shiffrin RM, Schneider W. Controlled and automatic human
information processing. II. Perceptual learning, automatic attending
and a general theory. Psychol Rev. 1977;84(2):127–90.
9. Brooke J. SUS: a “quick and dirty” usability scale. In: Jordan PW,
Thomas B, Weerdmeester BA, McClelland AL. Usability evaluation
in industry. London: Taylor and Francis; 1996.
10. Association for the Advancement of Medical Instrumentation;
American National Standards Institute. Human factors engineering:
design of medical devices. ANSI/AAMI HE75:2009. www.aami.
org/publications/standards/HE75_Ch16_Access_Board.pdf. Accessed
January 31, 2012.
11. Association for the Advancement of Medical Instrumentation;
American National Standards Institute. Medical devices: application
of usability engineering to medical devices. ANSI/AAMI/IEC
62366:2007.
12. Food and Drug Administration. Applying human factors and usability
engineering to optimize medical device design. http://www.fda.gov/
MedicalDevices/DeviceRegulationandGuidance/GuidanceDocuments/
ucm259748.htm. Accessed January 31, 2012.
13. Paes AH, Bakker A, Soe-Agnie CJ. Impact of dosage frequency on
patient compliance. Diabetes Care. 1997;20(10):1512–7.
14. Viller F, Guillemin F, Briançon S, Moum T, Suurmeijer T,
van den Heuvel W. Compliance to drug treatment of patients with
rheumatoid arthritis: a 3 year longitudinal study. J Rheumatol.
1999;26(10):2114–22.
15. Cramer JA, Gold DT, Silverman SL, Lewiecki EM. A systematic
review of persistence and compliance with bisphosphonates for
osteoporosis. Osteoporos Int. 2007;18(8):1023–31.
16. Zahn JD. Analysis: desirable attributes of insulin injection pens
that drive patient preference and compliance. J Diabetes Sci
Technol. 2011;5(5):1210–1.
17. Cambridge Consultants. Patients prescribe ease-of-use for the
medical device industry. http://www.cambridgeconsultants.com/news_
pr296.html. Accessed October 31, 2011.
2597
Author Affiliation:
Tandem Diabetes Care, San Diego, California
2609
Abbreviations: (FDA) Food and Drug Administration, (HF) human factors, (SUS) system usability scale
2628
Keywords: engineering psychology, human factors, human factors engineering, human factors research, usability
2643
Corresponding Author: Noel E. Schaeffer, Ph.D., Tandem Diabetes Care, 11045 Roselle St., Suite 200, San Diego, CA 92121; email address
noel.schaeffer at tandemdiabetes.com
2670
Figure 1. Early home screen design.
table with 4 columns and 7 rows
Table 1.Preliminary Analysis Phase
Study methodology
Sample description
Study focus
Outcomes
Web-based survey
350 patients, 125 clinicians
Focus on touch screen concept
Participants preferred touch screen interface
Web-based survey
45 patient insulin pumpers
Prioritization of information to be included on the graphical user interface home screen
Initiated design based on user feedback of home screen function priorities
Web-based survey
116 patients, 31 diabetes educators
Prioritization of all user tasks, including critical tasks, beyond the graphical user interface home screen
Used to select the architecture of the user task workflows
Focus group
40 patients
Focus on various options for terms and icons in the user interface workflow
Used to select specific terms and icons for user workflows that minimize confusion
Web-based survey
61 patients
Study color schemes used within the graphical user interface that maximize user recognition and minimize confusion
Selection of color scheme used throughout the graphical user interface
table end
2894
Figure 2. Design iteration of the home screen.
table with 4 columns and 9 rows
Table 2.Formative Evaluation Phase
Study methodology
Sample description
Study focus
Outcomes
Usability study
21 representative users
Workflow testing for completing critical tasks and for ease of use
Identified improvements required on critical tasks
Usability study
30 representative users
Focus on detailed critical task list and user errors and other HF
Usability errors identified and mitigated through changes to the graphical user interface
Usability study
33 representative users
Focus on test to ensure design changes made to the graphical-user-interface-mitigated errors
Design changes to increase task success rate
Usability study
15 representative users
Collect end user feedback on design changes made to user interface based on previous outcomes
Usability issues identified and design improvements implemented
Usability study
9 representative users
Graphical user interface navigation and presentation, software control usage and hardware control usage
Usability issues identified and design improvements implemented
Usability study
5 representative users
Graphical user interface navigation and presentation, software control usage and hardware control usage
Usability issues identified and design improvements implemented
Usability study
7 representative users
Graphical user interface navigation and presentation, software control usage and hardware control usage
Usability issues identified and design improvements implemented
table end
3176
Figure 3. Validated home screen design.
3185
Figure 4. Mean SUS score across all participants.